Chemistry · Lesson
Selectivity, sensitivity, detection limits and method validation
Assess performance for a defined use and matrix. Distinguish selectivity, response slope, detection probability and acceptable quantification rather than collapsing them into one sensitivity claim.
Back to the Chemistry mapBy the end, you can
- Distinguish response sensitivity from concentration-equivalent noise.
- Separate an observed decision threshold from a true detection concentration.
- Interpret a chosen ten-noise level without claiming a universal quantification limit.
- Solve a two-response toy and explain its selectivity boundary.
Validation asks whether the procedure suits the intended purpose
Method validation evaluates performance against a defined intended use, including relevant analytes, matrices, concentration interval and uncertainty or decision requirements. Verification checks that a specified procedure can be performed adequately under the actual laboratory conditions. An established method still needs evidence that it works in the setting where it is used.
Selectivity, precision, trueness, range, lower limits, robustness and uncertainty may matter differently for different objectives. Validation of the analytical stage alone may leave sampling and preparation questions unresolved. State what was assessed, under which conditions and with what acceptance criteria; a single attractive number cannot establish overall fitness.
A steeper response need not detect a smaller concentration
Sensitivity of a measuring system is change in indication per corresponding change in quantity; in a linear response model it is the slope. Detection performance also depends on noise, background, selectivity and the chosen decision probabilities. Sensitivity should not be used as a synonym for detection limit.
Two invented systems use the same response unit U and concentration basis. A has slope 2.00 U L mg⁻¹ and response standard deviation 0.100 U; B has slope 4.00 U L mg⁻¹ and response SD 0.400 U. Under a fixed linear additive-noise model, concentration-equivalent noise σ_C=σ_R/m is 0.0500 and 0.100 mg L⁻¹. B has the steeper slope but twice the concentration-equivalent noise.
This calculation is not complete detection-limit validation. Different response units cannot be compared by raw slope alone. Changing variance, interferents and uncertainty of the calibration parameters can require a different model.
A decision threshold and a detection concentration are different quantities
For an optional deeper original teaching model, one indication R at true concentration C follows a normal distribution with mean b+mC and known constant SD σ. Choose b=1.00 U, m=2.00 U L mg⁻¹ and σ=0.100 U. Calibration values and noise are treated as known; these are invented inputs, not method data.
Choose false-presence probability α=0.0500 and false-absence probability β=0.0500. The one-sided standard-normal 95th percentile is z=1.64485362695. A decision rule declares a positive indication when R exceeds R_crit=b+zσ≈1.16448536 U. A blank with C=0 has 5% probability of exceeding it.
The concentration corresponding to that threshold by inverse calibration is C_crit=(R_crit−b)/m≈0.0822426813 mg L⁻¹. At that true concentration, the mean indication equals the threshold, so only 50% of indications exceed it. It is not the true concentration giving 95% detection.
For the latter level under this model, C_D=(z_(1−α)+z_(1−β))σ/m≈0.164485363 mg L⁻¹. At C_D, 95% of indications exceed R_crit and 5% do not. R_crit is an observed-indication cutoff; C_D is a true-concentration performance level. Neither labels an unobserved sample as exactly zero.
The plotted probability is Φ((b+mC−R_crit)/σ), where Φ is the standard-normal cumulative distribution function. Estimated noise or calibration, changing variance, multiple observations and matrix interference require appropriate new treatment. The probability of a positive signal given a blank is also different from the probability that an observed positive sample is a blank.
Chance of positive indication
True toy concentration / mg L⁻¹
- Blank, threshold-equivalent level and detection level
- — Known-normal exceedance probability
Underlying data
| True toy concentration / mg L⁻¹ | Chance of positive indication |
|---|---|
| 0 | 0.050000000000000655 · Blank, threshold-equivalent level and detection level |
| 0.08224268134757362 | 0.5 · Blank, threshold-equivalent level and detection level |
| 0.16448536269514724 | 0.9499999999999993 · Blank, threshold-equivalent level and detection level |
Original known-normal, one-indication model with b=1.00 U, m=2.00 U L mg⁻¹, σ=0.100 U and fixed decision threshold R_crit≈1.16448536 U. Markers show true C=0, C_crit≈0.0822426813 and C_D≈0.164485363 mg L⁻¹, with exceedance probabilities 0.05, 0.50, 0.95. The curve is the stated normal-model probability, evaluated by numerical density integration. It supplies no observed sample result or validated method limit.
A chosen lower quantification level needs confirmation
In the same simple response model, choose a teaching concentration C_10=10σ/m=0.500 mg L⁻¹. Its net response above blank is 1.00 U, ten times the assigned response SD. The response-noise-only relative SD is 10%.
That result explains one commonly used scale, but it does not establish an actual quantification limit or acceptable total uncertainty. Required precision, trueness, selectivity and matrix behavior depend on the intended use. Lower-limit estimates need confirmation with appropriate evidence rather than automatic acceptance of a multiplier.
A second response can supply information only if its model is valid
Consider two invented exact response equations with no intercept or noise: R₁=2C_A+C_I and R₂=C_A+3C_I, with concentrations numerically in mg L⁻¹ and response coefficients in U L mg⁻¹. A is the intended analyte and I an interferent. At C_A=2.00 and C_I=1.00 mg L⁻¹, both indications are 5.00 U.
The coefficient matrix has determinant 2×3−1×1=5. It is invertible: C_A=(3R₁−R₂)/5=2.00 and C_I=(2R₂−R₁)/5=1.00 on the stated numerical basis. If only R₁ were used while falsely setting C_I=0, the result would be C_A=2.50 mg L⁻¹.
Two channels do not automatically prove selectivity, orthogonality or identity. The actual coefficients, interfering forms, conditioning and noise need validation. This nonsingular exact toy illustrates added information, without claiming experimental performance.
A performance claim should retain its tested boundaries
Report the entities, matrix, interval, conditions and procedure for each validation result. Robustness examines the effect of specified variations, while reproducibility comparisons state what locations, operators or systems changed. Neither term means every conceivable condition was tested.
If the target, preparation or matrix changes beyond that scope, assess whether further validation is needed. Fitness is a supported relationship between a procedure and its intended use, rather than a permanent label granted by one successful calibration.
Worked examples
Compare concentration-equivalent noise
Use systems A with slope 2.00/noise 0.100 and B with slope 4.00/noise 0.400 in the same units.
- A noise is 0.100/2.00=0.0500 mg L⁻¹.
- B noise is 0.400/4.00=0.100 mg L⁻¹.
- B has twice the slope and twice the concentration-equivalent noise.
Slope alone does not establish lower detection capability.
Distinguish the rule from its performance level
Use the invented known-normal model and α=β=0.05.
- Observed threshold is approximately 1.16448536 U.
- Its inverse-calibration concentration is approximately 0.0822426813 mg L⁻¹.
- At that true concentration detection probability is 0.50.
- True C_D≈0.164485363 mg L⁻¹ instead gives probability 0.95.
An observed decision cutoff and a true detection concentration are different quantities.
Interpret the chosen multiplier
Use σ=0.100 U and slope 2.00 U L mg⁻¹.
- C_10=10σ/m=0.500 mg L⁻¹.
- Net response is 1.00 U.
- Response-only relative SD is 0.100/1.00=10%.
A restricted noise-scale calculation, rather than a validated quantification limit.
Retain the interferent in the equations
Use R₁=R₂=5.00 U and the stated two-channel model.
- Determinant is 5.
- C_A=(15−5)/5=2.00 mg L⁻¹.
- C_I=(10−5)/5=1.00 mg L⁻¹.
- Ignoring I in R₁ gives 2.50 mg L⁻¹ instead.
Modelled added information separates the two chosen contributions in this exact toy.
Try it yourself
A new fixed linear toy has slope 5.00 U L mg⁻¹ and response SD 0.100 U. Find concentration-equivalent noise and the chosen ten-noise concentration.
Show the answer and reasoning
0.0200 and 0.200 mg L⁻¹.
Divide noise by slope, then multiply that concentration scale by ten.
In the known-normal model, keep slope and α/β fixed but double σ to 0.200 U. Find C_D.
Show the answer and reasoning
Approximately 0.328970725 mg L⁻¹.
C_D is proportional to the specified constant noise SD; the decision threshold also changes.
Does an indication failing the chosen positive rule prove zero analyte?
Show the answer and reasoning
No.
The model permits false absence and below-threshold indications from positive true concentrations.
Does merely adding a second channel prove experimental selectivity?
Show the answer and reasoning
No.
The coefficients, identities, interference scope, noise and model conditioning need evidence.
What to carry forward
- Validate for a stated purpose and scope.
- Separate response slope from concentration noise.
- Distinguish observed cutoff and true detection level.
- A chosen quantification multiplier needs confirmation.
- Multiple responses need valid models and selectivity evidence.
References
Teaching resources and reference material supporting this lesson’s explanations and examples.
- International Vocabulary of Metrology (VIM3)Joint Committee for Guides in Metrology (JCGM)SourceVIM3 4.12 sensitivity of a measuring systemVIM3 4.13 selectivity of a measuring systemVIM3 4.18 detection limitVIM3 2.24 reproducibility condition of measurement
- The Fitness for Purpose of Analytical Methods, third editionHelen Cantwell (editor); Eurachem Method Validation Working GroupSource · PDF
- Validation of Measurement Procedures that Include SamplingMichael H.Ramsey; Peter D.Rostron; Fernando C.Raposo (editors); JointEurachem/EUROLAB/CITAC/Nordtest/RSCAMC taskgroupSource · PDF
- Analytical Chemistry 2.1: 5.4 Linear Regression and Calibration CurvesDavid T.Harvey (DePauw University,USA)Source
- ICH Q2(R2): Validation of Analytical ProceduresICH Expert Working Group (Quality); International Council forHarmonisation; FDAhostSource
- Harmonized guidelines for single-laboratory validation of methods of analysisMichael Thompson; Stephen L.R.Ellison; Roger WoodSource