Engineering & Technology · Measurement, Units & Engineering Evidence
Test records, traceability and interpreting disagreement
One team reports a length of 50.02 mm and another reports 50.10 mm. Choosing the result with more decimal places will not resolve the disagreement. First establish whether they measured the same property of the same object under comparable conditions, and whether their stated uncertainties support the comparison.
The opening design chapter supplies the idea of a requirement and a test. This chapter explains its own measurement vocabulary and uses addition, multiplication, division, simple powers and averages. The final worked comparison introduces a square root; its meaning is explained beside the calculation. All dimensions, readings, devices and test results here are constructed teaching cases, not real measurements or equipment specifications. The mathematics measurement and data chapters offer optional additional practice.
Back to the chapterAfter reading, you can
Create reconstructable test records and investigate disagreement using comparable quantities, uncertainty and calibration evidence.
Leave a path from question to conclusion
A test record should let someone reconstruct what happened. Identify the purpose and acceptance rule, object and revision, quantity and measurement locations, operator and date, procedure version, instrument identity, relevant calibration record, environmental conditions, raw readings, corrections, calculations and uncertainty statement. Include unexpected events and the conclusion’s scope. Record enough detail for this test; a generic form with empty boxes is not evidence.
Preserve earlier versions when a calculation is corrected. Connect the new result to its original readings and explain the change. If the object or setup changes, give the new test a distinct identity. Otherwise an apparent improvement may be a comparison of different things with the same label.
Distinguish a record history from metrological traceability
Document or sample traceability follows the history of a file or object. Metrological traceability is a property of a measurement result: a documented, unbroken chain of calibrations connects it to a stated reference, and each calibration contributes uncertainty. An instrument’s serial number helps locate evidence; the number alone does not establish this chain.
A practical chain might connect a result to a working gauge, its calibration against a laboratory standard, and that standard’s calibration toward a realisation of an SI unit. The actual certificate relationships and uncertainties must support the claim. Calibration chains may branch when a result depends on several input quantities. “Traceable” is incomplete unless the reference and supporting evidence are clear.
Traceability does not guarantee fitness for a task. A well documented measurement with uncertainty of several millimetres may be unsuitable for checking a fit that changes by a tenth of a millimetre. The measurement need still determines whether the uncertainty is sufficiently small.
Follow a completed record back to its inputs
Here is a filled teaching record for the A/B comparison worked below. The object, procedure, people, times and calibration-input labels are invented additions to that case; no measurements or calibration certificates exist. The standard uncertainties are supplied model inputs, not estimates obtained from the single readings. The point is to make the path checkable and make missing evidence visible.
The measurand is the unloaded separation between the marked parallel faces F1 and F2 at centre-line C on panel P7 revision 2. Both records use procedure MP1 revision 1: support the panel horizontally without clamping, align the gauge contacts normal to those faces at C, bring them just into contact without visibly deforming the panel, and record the indication before applying the signed correction. Assume a stable 20 °C condition for this teaching comparison. Contact and environmental effects are included in the supplied total uncertainties; this is not a proposed real uncertainty budget.
This record lets a reader recover each corrected result and see that the rule concerns agreement between measurements, not whether the panel meets a fit requirement. In a real record, C-A and C-B would need the actual applicable calibration documents, correction functions, reference relationships and uncertainty contributions. Merely filling those labels does not establish metrological traceability. If a follow-up finds that B contacted an edge instead of C, retain R-B1, explain the changed measurand and give the corrected setup a new test record; do not silently replace the raw reading.
| Record field | A record | B record |
|---|---|---|
| Identity, person and time | R-A1; operator A; teaching day 1, 09:00 | R-B1; operator B; teaching day 1, 09:10 |
| Object, quantity and setup | P7 revision 2; F1–F2 separation at C; MP1 revision 1; stable 20 °C assumed | Same object and marks; unchanged; MP1 revision 1; same condition assumed |
| Instrument and calibration input | Gauge G-A; classroom input C-A supplies correction −0.02 mm at this indication | Gauge G-B; classroom input C-B supplies correction +0.03 mm at this indication |
| Raw indication retained | 50.04 mm | 50.07 mm |
| Correction and result | 50.04 + (−0.02) = 50.02 mm | 50.07 + 0.03 = 50.10 mm |
| Uncertainty statement | Total standard uncertainty uA = 0.02 mm, supplied | Total standard uncertainty uB = 0.03 mm, supplied |
| Comparison rule agreed in advance | Investigate if |B − A| exceeds 2uΔ; uncorrelated results assumed | Same rule; no dimensional acceptance limit specified |
| Event and conclusion record | No change reported in this model; retain R-A1 and R-B1 | Difference 0.08 mm exceeds 0.0721 mm; investigate, do not select a winning gauge |
Compare like with like before combining results
Check object identity and revision, locations, units, corrections, contact conditions, timing and environment. A maximum width and a centre-line width can both be measured carefully while answering different questions. If the measurands differ, averaging their values obscures the difference rather than resolving it.
If both results refer to the same stable measurand, compare their difference with the uncertainty of that difference. For uncorrelated results with standard uncertainties uA and uB, uΔ = √(uA² + uB²). The square root reverses squaring: √25 = 5. Shared references or conditions can correlate results, so this shortcut is not valid automatically. A comparison criterion should be chosen for the purpose before looking for a preferred outcome.
The diagram shows two reported expanded intervals and, separately, their difference against an illustrative compatibility threshold. Overlap of the two individual intervals is not a general statistical test. Incompatibility prompts investigation of changed quantities, missing effects or underestimated uncertainties; it does not identify which team is right.
An overlap that still needs investigation
Assume two independent, uncorrelated results on an unchanged object: A = 50.02 mm with standard uncertainty 0.02 mm, and B = 50.10 mm with standard uncertainty 0.03 mm. Their difference B − A is 0.08 mm. Its standard uncertainty is √(0.02² + 0.03²) mm = √0.0013 mm ≈ 0.0361 mm.
For this teaching comparison, choose a threshold of twice that standard uncertainty: about 0.0721 mm. The observed difference 0.08 mm exceeds it, so the results fail this chosen compatibility criterion. This is a prompt to investigate, not proof of a particular cause or a universal pass/fail rule.
Their individual k = 2 expanded intervals are A: 49.98–50.06 mm and B: 50.04–50.16 mm. They overlap, yet the difference exceeds the chosen threshold. Keep both records, check definitions and conditions, and plan a discriminating measurement. Do not average away the disagreement.
Try the reasoning
Does an instrument serial number establish metrological traceability?
Show the answer and reasoning
No. It identifies an instrument. The result needs a documented calibration chain to a stated reference, with contributions to uncertainty.
Why can two careful teams disagree without either making an arithmetic error?
Show the answer and reasoning
They may measure different locations or quantities, use different conditions or corrections, or omit an uncertainty contribution. The record helps distinguish these possibilities.
Why is the independent-results formula unsafe when both teams share an important calibration effect?
Show the answer and reasoning
The errors can be correlated. The uncertainty of the difference then depends on covariance, not just the separate squared uncertainties. Shared effects can cancel partly in a difference.
A record says only “Bracket B, 24.06 mm, gauge 17, passed.” The fit requirement concerns the maximum unloaded width, but the note may describe a centre-line reading. Repair the record enough to decide what should happen next; do not invent missing measurements.
Show the answer and reasoning
Retain the original note and mark the verdict unsupported pending clarification. Identify bracket revision and measured faces/location, operator/date, units, setup and procedure, raw indication, correction basis, instrument/calibration evidence, uncertainty and the actual fit limit/decision rule. Recover these from contemporaneous evidence where available; label unrecoverable items as unknown. A centre-line value cannot establish maximum width, so plan and identify a new measurement of the required quantity with an appropriate procedure. Record that result separately and explain how it addresses the requirement. Extra decimal places or a guessed correction cannot repair missing observations.
What to carry forward
A useful test result carries its method, history and uncertainty. Resolve disagreement by examining comparable evidence and the uncertainty of the difference.
References
Teaching resources and research supporting this lesson’s explanations and examples.
- International Vocabulary of Metrology (VIM3)Joint Committee for Guides in Metrology (JCGM)Reference guideInternational vocabulary of metrology (VIM3) — 2.6 Measurement procedureInternational vocabulary of metrology (VIM3) — 2.41 Metrological traceabilityInternational vocabulary of metrology (VIM3) — 2.47 Metrological compatibility of measurement results
National Institute of Standards and Technology (NIST)2 sources